Check that the vacuum pump oil of the lyophilizer works continuously, at least once a week. If the lyophilizer is short of oil, it shall be supplemented in time. Due to the frequent operation of air ballast, the oil consumption of lyophilizer is inevitable. Refer to the specific operating instructions of the vacuum pump for the make-up oil of the lyophilizer.
Before each operation of the freeze dryer, pay attention to removing the water in the cold trap chamber. Dry the cold trap if necessary. It is recommended to open the drain valve before each freeze-drying work, and then close it.
The lyophilizer shall check the liquefier at the bottom of the machine every few months for dust and dirt, and it shall be cleaned if necessary. When checking the heat exchanger, turn the chassis over to lie on the cushion. High-pressure air can be used to blow out the heat exchanger as appropriate. Dirty liquefier will increase power consumption and even damage the machine.

In case of an insufficient vacuum of the lyophilizer, first, check the drain valve and rubber sealing valve. It is recommended to remove the right side plate of the machine, then unscrew the drain valve, block the pipe orifice of the lyophilizer with a rubber plug, and then vacuum. If the vacuum required for operation can be reached, the problem lies in the sealing of the drain valve. The lyophilizer is generally caused by the attachment of drying residue and rag wire head. It is recommended to start the machine for vacuum pumping and blow off the attachments on the drain valve with airflow. If the above problems cannot be solved, the valve must be cleaned or replaced. Check the oil filling of the vacuum pump, replace the dirty oil of the lyophilizer if necessary, and check the pressure. Check whether the dry air grinding port is evenly coated with vacuum oil on the whole sealing surface. Vacuum the freeze dryer for inspection, and replace the vacuum rubber sealing valve in turn for inspection. Check the vacuum sensor to see if it is blocked by water evaporation residue. Replace the vacuum sensor if it is damaged.
In order to find out where the seal is not tight, the vacuum gauge can be directly connected to the suction port of the vacuum pump. If it can reach 0.011 mbar, it can be seen that there is no fault in the vacuum pump and measuring system. Sometimes the freezing temperature is not low enough to reach the vacuum. If the above steps can not determine the loose sealing, please check the small flange connection. Generally, these sealing rings do not require true air oil.